[Yaesu] FT-ONE - low output.
Danny Pease
dpease at adams.net
Wed Jun 1 14:37:29 EDT 2011
-----Original Message-----
From: yaesu-bounces at mailman.qth.net [mailto:yaesu-bounces at mailman.qth.net]
On Behalf Of Dr. Gerald N. Johnson
Sent: Wednesday, June 01, 2011 12:42 PM
To: Dr. David Kirkby
Cc: Yaesu
Subject: Re: [Yaesu] FT-ONE - low output.
On 6/1/2011 11:23 AM, Dr. David Kirkby wrote:
> On 05/29/11 12:49 PM, Rob Atkinson wrote:
>> I don't argue about the Bird accuracy but if you don't like a 43, how
>> will
>> you be content with any hammy hambone watt meter out there marketed to
>> hams?
>> I think my point is that hams don't really require great accuracy in
> power measurement.
>> Unfortunately I've never tested any ham meters in a lab, so I don't know
> how good/bad they are.
Then how do you KNOW some Birds won't do 10% accuracy?
Some copies of the 1956 vintage Monimatch are often sold for CB. Near
all including Bird and Collins wattmeters have an adjustable resistor
for calibration. Bird pots are hidden in the slugs generally. Many CB
SWR bridges with watts indications have a serious sensitivity to applied
frequency.
>>> If you want to achieve accuracy on a budget, consider getting a
>> thermocouple
>> RF amp meter (they show up at hamfests for a few dollars/pounds/euros but
>> are often hidden under junk because most hams don't give them much value
>> which is good for you) that measures 6 A and a good 50 ohm non-reactive
>> dummy load. Mount the meter in an aluminum box in series with UHF jacks
>> mounted on the sides. Measure your current into 50 ohms and calculate the
>> power. Actually with your legal limit a 2 or 3 A meter is fine.
>> I don't have much clue about how accurate that would be. I'm not an
> expert in RF power measurement, but when it comes to Bird meters, I
> probably know more than 99.99 % of hams. At least one review on eHam
> describes the meter as "lab grade". The reviewer is seriously mistaken.
>> I think my main issue with Bird is the claimed accuracy and the cost. If
> the accuracy was as good as Bird state, then I'd have no problem with
> the cost.
>> But when Bird claim +/- 5% of FSD, and I know some don't even meet 10%
> of FSD, then I do have a problem with the meter.
The slugs are destructible and can be misapplied with the wrong line or
meter. Especially low power slugs. Coupler based wattmeters are limited
in their directivity. Using a lower power slug for reflected power tends
to destroy the low power slug and to contribute to considerable error
because even with a perfect load the 50 watt slug set for reversed power
seeing a KW forward power senses that forward power and with directivity
about 30 dB shows 1 watt reflected for every forward watt. The lower
power slug has tighter coupling to the line and so when applied for more
sensitive reflected power can lead to excess dissipation in the coupled
line termination and detector leading to damage of the slug. That's a
destructive misapplication and why I no longer consider purchasing low
power slugs at a hamfest or epay.
>> I've often wondered how difficult it would be for hams to make a water
> calorimeter. For high power, that's how its measured in a standards lab.
> Of course, this would require that you transmit for a time sufficient to
> reach thermal equilibrium, which would exceed the duty cycle of most ham
> rigs. Making one would be a major project, but an interesting one and
> within the capabilities of a dedicated constructor.
There are two schemes for water calorimetry. One used occasionally for
primary standards uses controlled water flow through a well insulated
load and measures the temperature rise of that water flow.
A better ham solution uses the load in a well insulated water volume.
With that water stirred to keep the temperature uniform, its easy enough
to compute the rate of temperature rise by the power applied. 3.412 BTU
per watt. I believe the fundamental is that 1 pound of water rises one
degree F for each applied BTU, though that may off by a factor of a
thousand. However its easy enough to measure the rise in water bath
temperature vs time with applied RF power. Then to let the water cool,
and apply easily measured DC or power line frequency AC power to the
same load and adjust for the same rate of water bath temperature rise.
The insulation can be a 3ドル foam picnic cooler, the load can be a coil of
small diameter coax immersed in the water in the cooler. A rapid
response dial thermometer cost me 7ドル in the hardware store, should I
want to do that I keep a couple on hand for checking my vehicle's AC
performance and when farming for checking soil temperature when planning
for planting. By calibrating with DC, the thermal mass of the load and
the water is combined so the experimenter need not measure the volume of
the water precisely. But the technique gets RF power measurement down to
the fundamentals of measuring time and temperature. And with the DC
calibration the precision is limited mostly by the operator care and the
calibration of the DC instruments which can be traced to primary world
standards if need be.
>> I think for me a spectrum analyser is the easiest solution. It may not
> be the best, but is good enough. I could calibrate a ham meter against
> that.
>Being that the spectrum analyzer is essentially a logarithmic scanning
receiver with tolerances of a few dB, I'd not consider that precise
enough for a calibration standard. I much prefer a good (TEK or
HP/AGILENT only) oscilloscope. I own a Tek 475, the 465 is great for HF,
like its direct competition the HP1740. I'm less of a fan of digital
scopes, but the makers don't build analogue scopes anymore.
>73, Jerry, K0CQ
Bird 43 meters are only spec'ed at full scale, anything else and their
spec's are out the window. Beyond that, the weak and variable link is the
element itself, which is where the real inaccuracy lies. Unless you have
access to a calibrated power meter, such as a calorimeter, you never know
what the accuracy of a Bird element is at less than full scale. Using a 2500
watt element to measure 1500 watts is very common, but how many have the
element calibrated at 1500 watts? Not only that, is your element calibrated
at every ham band you use? Have you ever applied too much power to a lower
power element? Have you ever dropped an element? Have you ever used a low
power element for reflected reading and exceeded the 10:1 ratio. I know from
experience it is very easy to damage a Bird element by using, say a 5 watt
element in the reverse but run 100 watts through the meter and watch the
reflected reading creep up. That creeping is heat causing the diode to leak
and/or the load resistor overheating, both affecting calibration. If you
antenna system (load) is not a non-reactive 50 ohms, Bird 43 elements and
meters accuracy are again suspect in accuracy.
It is very easy to know your meter is accurate and any one power level and
frequency by comparing to a known good meter (not one you think it accurate)
but it is much tougher to know a meter is accurate over a large range of
frequency and power levels. Measuring PEP is basically a guess unless you
use an accurate Oscilloscope and a really good dummy load. These "peak
reading" meters do not measure PEP, they just add a capacitor to hold the
meter deflection closer to the top of the meter swing. It is much cheaper
and much more accurate to use an oscilloscope and a dummy load/attenuator to
measure power than to buy a really good wattmeter. I have also seen a few
big name oscilloscopes that were not very accurate when measuring radio
frequencies, so again, you are back to having to have a known good reference
to know where you are.
One place I worked at with its own calibration lab calibrated Bird 43
elements against a Bird standard at the power levels and frequencies we
tested our transmitters at and made up a cheat sheet for each element
showing what fudge factor to add or subtract from the meter reading, power
vs. frequency. This was a lot of work for the calibration techs, so we
finally went to Bird 4420 meters (+/- 3% at all power levels within the
frequency range of the sensor, in our case 25 MHz to 1000 MHz), that solved
almost all of our accuracy issues.
Making your own thermal meter is really pretty cheap, just not convenient to
use for most of our applications.
All that said, 10% is more than accurate enough for amateur use.
More information about the Yaesu
mailing list